100%A-35%A over 40' where A=15% ACN/0.1% TFA, B=50% ACN in 0.1% TFA.
The first of 3 columns ordered failed a tailing factor spec. for the main peak right out of the box but gave otherwise good, if uniformly earlier (than a previous lot used in method development) peaks. The second column was better with respect to peak shape and was a better match to the first lot with respect to retention times but it seems to have died after having run just 3 sample sets. It was stored in ~75% ACN/water for a few weeks between its last sample set and its recently failed attempt to run more. Another XBridge that was abused in storage (was stored in ~50%organic/acetate buffer for 3-4 months) was utterly useless for this particular assay.
Waters states a preference for ACN as the storage solvent for this column but they are not adamant about it.
My questions:
Have others had similar experiences?
Is this particular phase *that* sensitive (damaged by long term storage in anything but straight ACN)?
Is Waters addressing lot to lot variation on this packing?
DR