Quadrupole 5973 cleaning procedure

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

4 posts Page 1 of 1
Hello!
I need to clean my quadrupole assembly on my agilent 5973.
Do you have a procedure to clean it?
I know that i can't use ultrasonic bath
Thank you!

Hi Rick,

Do you mean that you want to clean your ion source assembly? If so, then you can incorporate an ultrasonic cleaning stage on the metal parts but not on the insulators, source heater assembly or filaments. The procedure is in the manual and can be followed quite easily. You should also have the cleaning accessories kit that came with the MS.

Regards,

Ralph

Just remember that Agilent no longer recommends a drying oven bake out step at the end of the cleaning procedure. They say you should let it air dry at room temp, not that I've noticed any real difference either way.

Rick,

The Agilent's engineer told me that MSD needs to be autotuned when it's in trouble. The ion source may become dirtier everytime when autotune is executed. However, there is no fixed frequency as suggested by him. I recommend tuning the MSD every month to track the change of MSD performance.
4 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Google [Bot], Semrush [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry