Have replaced practically everything at this point (septum, liner, o-ring, ferrules, gold seal, even the column).
Still, the leak persists.
I'm pretty confident that the source of the leak is in the inlet, but I'm unable to be 100% sure because we don't have a helium detector on hand (I'm trying to borrow one for a day or two).
When I use the Service Mode Front Inlet Leak Test (10 psi), it passes the test every time because there is no pressure drop from Inlet to Column.... but instead, I'm seeing an increase from inlet to column.
The machine is measuring total inlet flow at approximately 0.8 ml/min of helium, and the total column flow at ~0.99 ml/min of helium.
Does anyone have any ideas why this might be the case?
Have I missed something obvious?
Thanks in advance.