Tocopherol & Tocotrienol Part II

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
Hi everybody

Having used this forum extensivly I feel it’s time to give a little back. For once I don’t only have questions but I'm posting a (almost) complete method :) . With inspiration in ljc’s method in post “Tocopherol & Tocotrienol by HPLCâ€
Kind regards
Leadazide

HI, Leadazide,
Thanks for the sharing.
It is interested to me because these are very hydrophobic molecules/isomers I never work on.
Do you the following answers?
what is the retention time if in a C18 column with mobile phase of 98% ACN or THF or IPA?
what if it's an endcapped RP column?
Do you know if the DIOL is better than CN or Amino due the different column stationary chemistry? or you/they happened to start from DIOL?
Do you know anything other reason of choosing DIOL?
what if in a straight silica gel column?
Thanks and have more funs,
Excel

These compounds were separated nicely on our
Unison UK-Silica column:

http://www.silvertonesciences.com/modul ... TI166E.pdf

Bryan: Darn, that seperation looks sweet!! :)

ym3142: On the RP system we use a ODS hypersil 125mm x 4.6mm column with methanol:water 96:4 at 1 ml/min. This gives us a rt of 7 for delta, 9.5 for gamma + beta and 11 for alpha. Right now our system that runs the RP method is occupied with other methods, so I can't check with your specific eluent in the near future.. sorry:(

The reason we use the DIOL is that i read several different examples of methods which used this column. I had tried with both CN, amino and a silia column.. I didn't get any proper seperation with these, but it might very well be the columns that are ruined since they were just lying around and are at least 10+ years old. So when I had to order a new column anyway I ordered the DIOL, since I had seen example chromotograms with this and liked what i saw plus we didn't have a DIOL column and I like to have lots of options lying around in case we have to try something new in the future.. :)
Kind regards
Leadazide

If were me, I will try some of C18, C8, Phenyl columns; then, their endcapped(hypersil is). I am sure one of these column will be good.
though you have try different organic and mixture, buffer, addictives, temperature, gradient, even PH. good luck.
Excel

I recommend C30 RPcolumn for separation of tocopherols. It can separate gamma and beta tocopherol at 15 C completely. You can see its separation on the following website:

http://develosil.net/eappli019.htm
6 posts Page 1 of 1

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry