- Posts: 2
- Joined: Mon Jun 05, 2017 9:31 am
I am hoping that someone on the forum might be able to shed some light on our injector issue. We had a COC injector port installed on the back inlet of our 5970a GC - FID.
When the autosampler (G4513 plugged into back inlet port) lowers to inject the sample - the A/S stops with a shudder and goes into fault mode - the plunger remains depressed. I can manually inject the sample at this point and then raise the injector. Needle is not bent.
I have tried realigning the A/S using the automatic align on the turret - and all seems OK. I can manually inject a sample, so I am happy with the alignment of the 0.53um guard column and associated COC parts and we have the 0.53 on column nut with septa in place. I have also changed the needle guide to the on column needle guide that came with the kit.
The needle is an 5ul Agilent on column syringe (5182 0836) with 0.32 needle, inlet pressure is 10psi has a septum purge of 15ml/min. These pressure were set by the Agilent engineer so I am not sure whether these are what we should be using - but thought I would include them if it help narrow down the A/S issue.
Many thanks for any help on this issue,
Matt