Series of GCs in two different laboratories (mixture of 6890's and 7890's).
Some GCs in some laboratories are able to successfully run a method, while others are not.
There are no problems with blank injections or standard injections. There is never a problem with the first sample injection.
All of the trouble starts after the first sample injection. The problem is a co-eluting peak with the analyte of interest.
I work in an FDA regulated environment and this is a fully validated analytical method that has been in use for over 10 years. We have transferred this method literally 15-20 times and have never observed this problem.
In an effort to troubleshoot I have:
1. Taken column, liner, syringe, septa, and sample preparations from a non-working GC to a working GC and the working GC has no problem.
2. Done the reverse of #1 and the non-working GC has the problem.
3. Have tried different columns, liner, syringes, septa. The analysis works on the good GCs, not on the bad.
4. Replaced inlet, gold seal, FID, split vent copper tubing, and split vent trap on bad GC. Problem presisted, but the interfering peak was about 1/10 of the normal size.
5. In light of #4 suspected contamination of sample contact surfaces within the inlet, so I replaced the reducing nut from the inlet to the column. This caused my problem to regress back to original size.
6. Replaced all carrier gas traps.
7. Replaced FID jet and cleaned ceramics of detector.
As I said, this happens arcoss GCs (6890/7890) and laboratories. I have verified all instrument methods are the same and all relevant parts in the GC are of the same part number.
The question is: what am I missing? What haven't I looked at?