If we still assume adsorption and that the cause is your column and not your injection port then things to try are (in no particular order)
1. New column
2. Take 2 metres off your column - in case it has contamination at the front end.
3. Inject 1000ppm (at 50:1 split) between each calibration to saturate any sites. It can work as long as you just check first that there isn't carryover.
4. Another trick that I have used( but it depends on the analysis) is to add another component to your sample at quite a high level. This component is selected to contain a functional group and acts in a sacrificial way to bind adsorption sites.
1. I will test another column as soon as possible. I think I will test another manufacturer juste to make a comparaison
2. I'll try this too
3. For my calibration, I use a sample which consist of 26 products (24 allergen and 2 IS). Maybe 10 of them have some hydroxy groups. Which products can I use to saturate the actives sites? For the carry over : between each standards we make a blank, and we never see some strange peaks in this blank run.
4. I think it's the more easy solution to test, as I said in point 3 I already use a quite complex sample, so what kind of coumpound may I use to make this saturation?
For the determination of symmetry at 10% of the peak :
With the apolar column:1.20
With the polar one :1.11
And I repeat that nothing change apart the column, and the day of the experiment
@ chromatographer1 : What can I do against this kind of thing ? a lower injection temperature? desactivation of this active sites?