I know this is a popular topic but I'm looking for opinions on the best way to maintain check valves using a UPLC system. We experience a lot of problems with check valves sticking, causing erratic pressure etc but one of our engineers recommend to me to flush the A1/B1 lines in 90:10 Aqueous:Acetonitrile for about 30 mins following a long gradient run using buffered mobile phase?

I did this and for 10 months now not a single issue with the check valve, do you think its just luck or is there something to priming out the A1/B1 lines to wet the seals and flush away and residual buffer? There seems to be dozens of tips on how to stop check valve failure!
Warm water it the best flushing solvent however you should renew it at lest once a day to prevent algae growth. 5-40 % i-propanol/methanol/acetonitrile should be fine too.
Best regards,
Dmitriy A. Perlow
Flushing post run is always a good idea.

Also, if one of your MP is straight ACN, see if ceramic check valves are available. ACN can cause problems w/ the more traditional ruby ball/sapphire seat.
Thanks,
DR
Image
If more of my customers would listen and follow advice about a post-run flush method I could die a happy man.. :)
Kind regards
Leadazide
I agree with DR, after acetonitrile and/or buffer mobile phase/s have been used for a project, it is useful to flush the system with 85% water / 15%methanol. When buffers with a very high salt content are used, it is also useful to back-flush the pump pistons, please see the pump’s user manual.
Kind regards,
Ade Kujore
Marketing
Cecil Instruments
Cambridge
United Kingdom

email:- ade.kujore@cecilinstruments.com
telephone:- +44 (0) 1223 420821
web site:- www.cecilinstruments.com
Registered Number 909536
I agree with DR, after acetonitrile and/or buffer mobile phase/s have been used for a project, it is useful to flush the system with 85% water / 15%methanol. When buffers with a very high salt content are used, it is also useful to back-flush the pump pistons, please see the pump’s user manual.
Definitely a flush with buffer free mobile phase after. Another thing I have done, especially if the instrument will be used within a day or two after is to keep a very slow flow through the instrument of 50/50 water/methanol, about 0.2ml/min for non uplc systems. This seems to keep the check valves clean and also keeps the piston seals from sticking to the piston, which will cause wear on both once flow resumes. The pressure is normally low enough that it doesn't cause excessive wear on the pump seals and you don't worry about any microbial growth in the system.
The past is there to guide us into the future, not to dwell in.
6 posts Page 1 of 1

Who is online

In total there are 410 users online :: 0 registered, 0 hidden and 410 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 410 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry