- Posts: 27
- Joined: Tue Aug 15, 2006 11:11 pm
I'm working with the Agilent 1100 HPLC system in conjunction with the Micromass Quattro LC system. Following tuning of our test compounds, during the running of samples we often see baseline elevation/interference at the column switching timepoints (ON to MS/MS -> baseline jumps). Quantitation of these compounds still work but the signal to noise performance is extremely poor because the baseline jump (caused by excess tuning solution that I suspect is still in the MS system somewhere) interferes with the peak.
After the column switch is OFF the baseline returns to normal.
I've tried thoroughly cleaning what I can readily get to (cone, needle) but I can still see the existence of both parent and daughter ions on the tune page, even when the HPLC unit is OFF.
I've reduced the concentration of the tuning solutions fairly significantly to avoid this problem but it still significantly affects my samples near the LLOQ of the assay.
Thanks in advance,