Hi all,

I am encounter bfb evaluation for gcms fail on ion 75. The probability may be is ion source dirty. However we just finish ion source cleaning and without running any sample. May i know have any other probability beside that?

Thanks
Seet
After you re-installed the newly cleaned source and pumped the MS down-- Did you perform an Air and Water Check? I assume you are using Autotune for your BFB. Are you able to get a GOODTUNE.SAV if you manually choose/average your scans?
After you re-installed the newly cleaned source and pumped the MS down-- Did you perform an Air and Water Check? I assume you are using Autotune for your BFB. Are you able to get a GOODTUNE.SAV if you manually choose/average your scans?
In addition to this: you could copy/paste the checktune/autotune report here.

Most likely something went wrong in re-installing the source or pumping down.
3 posts Page 1 of 1

Who is online

In total there are 29 users online :: 2 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry