mobile phase made of Ammoniumacetate buffer
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
9 posts
Page 1 of 1
I am using a mobile phase made of ammonium acetate buffer and acetonitrile. I found the standard peak area count decreased with the age of my mobile phase. The peak area cound went back when I changed to a freshly prepare mobile phase. Anyone experienced this? how to explain it?
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
What is the pH of the buffer? How do you degas the mobile phase?
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
Some compounds have fairly strong changes in absorbtion coefficient as a function of pH and/or organic content. Phenolics are a good example. Usually this effect is accompanied by retention shifts when the average ionization of the molecule changes.
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 9
- Joined: Mon Jul 31, 2006 4:11 pm
Thanks.
I also experinced this before when I worked in my prevous job. The buffer used is pH 9 and detected at 210 nm. The run can not be more 24 hr long because the bracketing std always failed the 3% recovery limit.
The current method using buffer at about 6. I checked the mobile phase pH. It didn't change. I degassed the mobile phase by filtering through 0.45um filter. No additional degas performed.
I also noticed, the peak area decreased much faster when I used a bigger container for mobile phase.
I also experinced this before when I worked in my prevous job. The buffer used is pH 9 and detected at 210 nm. The run can not be more 24 hr long because the bracketing std always failed the 3% recovery limit.
The current method using buffer at about 6. I checked the mobile phase pH. It didn't change. I degassed the mobile phase by filtering through 0.45um filter. No additional degas performed.
I also noticed, the peak area decreased much faster when I used a bigger container for mobile phase.
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
At pH 9, you could have picked up CO2 from the air, which then could change the pH and therefore the absorption. This would happen faster in a larger stirred vessel. It is still possible at pH 6, but somewhat less likely. On the other hand, ammonium acetate is not a buffer at pH 6...
-
- Posts: 308
- Joined: Tue Aug 31, 2004 12:26 pm
Maybe you are losing ammonia from your buffer at pH 9. Why not measure the pH of the fresh mobile phase and then the mobile phase after 24hr of use under your normal conditions.
Measure the pH after the addition of the organic solvent. Measure it actually in your reservoir bottle when you only have a small amount left and a big headspace in the bottle. Since you are looking for a change, the interpretation of the absolute figures is of less importance.
Measure the pH after the addition of the organic solvent. Measure it actually in your reservoir bottle when you only have a small amount left and a big headspace in the bottle. Since you are looking for a change, the interpretation of the absolute figures is of less importance.
-
- Posts: 9
- Joined: Mon Jul 31, 2006 4:11 pm
Thanks a lot. these are the good explainations. I will perform your suggestions.
9 posts
Page 1 of 1
Who is online
In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 21 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 21 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science