My problem was that I was getting increased pressure and a quick deterioration of peak shape after about 10 sample injections. I assumed this was due to the HPMC
I changed the method to precipitate-out the HPMC (in Acetonitrile) prior to analysis - stock (and taking an aliquot of the clear upper soln and diluting it with Buffer). My final organic conc in the MP ended up as 5% (to get it lower, ie 2% I would have to add more sample to make the stock at which point the active starts to precipitate (not freely sol in ACN). With the higher org content, I get a dip that interferes with a peak of interest.
So it looks like I have to inject the sample diluted in buffer/2% ACN with HPMC present.
I added a guard column, Gemini C18: 4.0mm (L) x 3.0mm (ID), and this seems to work for the most part. The peak-shapes are good afer 20 or 30 samples, but the pressure begins to build (in the column, not the guard - I checked for that).
I have been covertly (Scientists in our company generally does not do this) taking the columns apart, replacing the top mm of packing with fresh packing, and sonicating the frits in nitric acid. This works fine.
OK, that was a bit long winded, but my question is how can I prevent this rise in pressure? I do not want to change the sample prep - Is there a better kind of guard column that I can use?
Thanks