I am running GPC with a Waters Ultrahydrogel column and can not seem to get rid of a double peak that shows up near the end of my runs. The peak shows up on my refractive index detector but not my fluorescence detector. My eluent is phosphate buffer. I make all my samples in the same eluent buffer. The Vo for my column is ~6min. and a Vt of ~15min. The flow rate is 0.8ml/min and the eluent sparged with helium almost continuously.
On RI detector, here are some findings:
(1) Peak is a positive double peak when running with a blank (i.e., elution buffer)
(2) Peak is a negative double peak when running a 1 to 10 dilution of a FITC labelled sample
(3) Peak is a positive double peak when running non-diluted version of FITC labelled sample
(4) During equilibration mode, there is no peak that shows up at all.
**I do get a valid peak on fluorescence detector but don't know how to interpret this ghost peak on RI detector.
I have been flushing my manual injector with water in between sample runs using a needle port cleaner so I am pretty sure it is not contamination between samples. Could the refractive index detector cell be dirty? If it was, why don't I see a ghost peak while in equilibration mode? Also, I don't see how this can be a solvent peak b/c it shows up near Vt and also, it is quite a significant peak.
Any ideas for what this is or how to get rid of it would be helpful.