GC-MS Masshunter - inserting dilution after sequence has run

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

4 posts Page 1 of 1
Hello,

Anyone know if it is possible to amend the dilution factor in GC-MS Masshunter Quant software after the sample has already been run? As I omitted the dilution factor from my sequence list when the run was initially started, it is calculating sample results using a default dilution factor of 1 currently.

Thanks
Aaron
Hello

1.Load your sequence
2.In sequence table change dilution factors
3.Save sequence
4.Click "run sequence" button
5.In "Start Sequence" dialog choose "reprocessing only"
6.Run sequnce

It will run sequence with corrected dilution factors
Check report if you can see correct dilution factor value

Good luck

Regards

Tomasz Kubowicz
Not sure about your version, but in mine you can just right-click anywhere on the data in the quantitative software, add column => "Dil." and put your factor there. Re-analyze and done.

We usually never put dilution factors in the sequence table.
Thanks for the replies folks - got it sorted with the extra column in the quant table.
4 posts Page 1 of 1

Who is online

In total there are 331 users online :: 0 registered, 0 hidden and 331 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 331 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry