I recently started analyzing a new compound (to us) and found some developed methods in literature. After testing a few of the methods I decided on one that seemed to give good peaks and a workable time scale (ie the retention time was neither too soon nor too long).
The mobile phase consists of water + 1% TEA, adjusted to pH 2.4 by phosphoric acid. I then mix the water with methanol 65/35 v/v (but measured gravimetrically). The pH is then re-adjusted to 2.4.
Having made three batches of mobile phase, it seems that the retention time is extremely sensitive to pH. It appears that even a very small change in pH (difference of 0.05) can result in the retention time shifting from 7.5min to 6.5min.
Is it common to experience this type of behaviour? Should the mobile phase be controlled within such a tight window? Or perhaps a standard can be used to adjust the retention time at the start of each day or whenever new mobile phase is added to the system.
I had also tried using a phosphate buffer with water and methanol, but the peak shape (tailing) wasn't as good as the above mentioned mobile phase.
Regards