Waters columns on Dionex Ultimate 3000

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hello,
I encountered an interesting problem when using Waters columns on Dionex Ultimate 3000.
Applying original analytical methods developed on Waters HPLC systems with Waters columns on Dionex HPLC I observed unacceptably poor peak shapes.
By changing the columns with similar ones from other types like Hypesil or Inertsil the peak shape improves.
The first thing I suspected was the improper fitting because of a specific column design.
However, changing the sample solution with mobile phase considerably improved the peak shape even by using the Waters column.

Has anyone had any experience with such problems?
Does anyone have some idea what might be the reason and how this problem can be overcome so that I can use the original methods on Dionex by using original Waters columns?

Thank you in advance!
Boyan Yordanov
If a change to your sample diluent resulted in better peak shape on any system, chances are that your original diluent was too strong for your A phase and that it worked on any particular LC was more or less just dumb luck. Stick to keeping your sample diluent weak relative to your initial mobile phase composition (this means A phase % organic > that of your sample diluent for reverse phase).
Thanks,
DR
Image
yes problem known to me too.

especially when the method was developed on an Alliance system (or older 717+ Injector).
These types of injectors are (by design) quite "robust" with injection of "bigger" or let's say normal volumes (10 / 20 µL) of high organic solvents (eg. like reference substance in pure methanol), even if the initial composition of mobile phase is high in water.

When transferring such methods to an Agilent 1200 system, we also observed peak distortion of the reference substance.
Lowering the organic content in the injection solution solved the problem.
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry