Peter, I have to correct you on this one. I can show plenty of examples where the deuterated molecule elutes earlier than the non-deuterated component in GCMS.
It would indeed seem logical that the increased molecular mass makes it elute later. But deuterated compounds, compared to the same non-deuterated molecules show less affinity towards the stationary phase. If you combine both effects, it elutes earlier
The deuterated molecule will ALWAYS reach the detector first in a quadrupole.
Peter, the units are minutes. So in my example, THC-d3 has an RT of 1.22 minutes, THC is at 1.23 minutes, but my THC-d3 2 ion might be 1.22 minutes and my THC 2 ion might also be 1.22 minutes. The XIC shows all of the peaks essentially on top of one another but there is a slight difference.
I know that the deuterated analyte will always come first due to reaching the detector first (this is always true in a quadrupole, for GC/MS and LC/MS/MS). It's one of our fundamental criteria for calling a positive result.
I'll see if I can redact a chromatogram and upload a picture to better see what I'm talking about.