What can I expect for column lifetime when doing rapid resolution chromatography with balistic gradients on a 2.1*20 mm Sub-2µm reversed phase column? My mobile phases will be Acetonitrile and Water both with 0.1% Formic acid. How many injections can one make before the column starts to deteriorate?
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
That depends primarily on the quality of your sample prep technique. I have seen over 2000 injections after a solid protein precipitation, but I am more comfortable to bet on 500+ injections. It also may depend on the quality of your gradient procedure.
-
- Posts: 13
- Joined: Wed Jan 11, 2006 2:29 am
I have some experience in High-through put LC/MS for library chemistry, we use ACN and water with 0.01% TFA as MP, and use 2.1*50mm,5um column. In average we could ran about 10,000 injections for one column. But we also use MP with 1mM NH4COOH in ACN and water, it only ran about 5000 injections.
We have 5 LC/MS, and each LC/MS ran almost 24*7 Hr per week.
We have 5 LC/MS, and each LC/MS ran almost 24*7 Hr per week.
The more time I have to think;
The more time I have;
The more time I have;
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Sorry, I had assumed that you are dealing with a sample of biological origin, such as a plasma sample. For any sample, column lifetime depends on the amount of junk one puts on the column. For very clean samples and clean mobile phases, columns can last much longer.
A long time ago, I demonstrated that a column does not even change in 10 000 injections, if it is protected by guard columns.
A long time ago, I demonstrated that a column does not even change in 10 000 injections, if it is protected by guard columns.
-
- Posts: 15
- Joined: Thu Oct 13, 2005 11:01 pm
Also depends on which column you use. At the 2005 BMSS meeting there was a presentation from GSK covering some work they'd done with sub 2 micron columns. They noted a difference in lifetime between different manufacturers columns. Examples were 1 brand lasting only ~400 injections while others were ok after 2000.
-
- Posts: 216
- Joined: Mon Apr 04, 2005 12:26 am
I did a lot of plasma analyses with protein precipitation. See
http://users.chartertn.net/slittle/default.htm
first topic, the text.pdf document
Initially we used a guard column very successfully getting 200 to 400 injections before changing the guard column. However, after a while we noticed that the guard columns being supplied were not as high a quality. I would test the guard column individually and found that it significantly decreased the quality of the separation. Thus, I then started to just use the column alone.
I then could get 400-800 injections before the quality of the separation decreased. I often changed (usually every 200 injections the 0.25 micron filter before the column if the backpressure increased significantly.
Often I could clean the column (10 or so column volumes) by backflushing with water/methanol (90/10) containing 1% formic followed by water/methanol (5/95) containing 1% formic, then just pure methanol. The columns would then often be OK.
A lot of our columns used in routine analysis fail due to compression of the bed. You can take the frit off the front of the column and see the compression. We have slurried some stationary phase and added it to the front of the column and brought the columns back to their initial performace spec. Sometimes if the column is very discolored at the head of the bed, we scrape out a little of the bed and then top it off with some more slurry, and carry it through the washing.
Depends on how much you value your time. Most people consider columns disposable after a reasonable number of injections. Looking at the mode of failure might give you insight into selection of a more robust column..
http://users.chartertn.net/slittle/default.htm
first topic, the text.pdf document
Initially we used a guard column very successfully getting 200 to 400 injections before changing the guard column. However, after a while we noticed that the guard columns being supplied were not as high a quality. I would test the guard column individually and found that it significantly decreased the quality of the separation. Thus, I then started to just use the column alone.
I then could get 400-800 injections before the quality of the separation decreased. I often changed (usually every 200 injections the 0.25 micron filter before the column if the backpressure increased significantly.
Often I could clean the column (10 or so column volumes) by backflushing with water/methanol (90/10) containing 1% formic followed by water/methanol (5/95) containing 1% formic, then just pure methanol. The columns would then often be OK.
A lot of our columns used in routine analysis fail due to compression of the bed. You can take the frit off the front of the column and see the compression. We have slurried some stationary phase and added it to the front of the column and brought the columns back to their initial performace spec. Sometimes if the column is very discolored at the head of the bed, we scrape out a little of the bed and then top it off with some more slurry, and carry it through the washing.
Depends on how much you value your time. Most people consider columns disposable after a reasonable number of injections. Looking at the mode of failure might give you insight into selection of a more robust column..
Sailor
6 posts
Page 1 of 1
Who is online
In total there are 221 users online :: 1 registered, 0 hidden and 220 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: Semrush [Bot] and 220 guests
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: Semrush [Bot] and 220 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science