Agilent 6890 and Attenuation

Posted: Wed Jun 21, 2006 11:15 am
by Anthony Doyle
Recently one of our analysts suggested to increase attenuation, in order to improve sensitivity for the analysis of one residual solvent. We injected a standard, with the following 'table' inserted into the 'runtime' option in the 'instrument actuals' part of the method.

Time Parameter Value
0 Attenuation 0
20 Attenuation 5
22 Attenuation 0

No increase/decrease in area was observed for the two peaks that elute between 20 and 22minutes.

Does anybody know anything about altering attenuation with 6890's?! Any pointers would be really appreciated.

Addendum

Posted: Wed Jun 21, 2006 2:19 pm
by Anthony Doyle
Apologies

We're using an FID on the 6890N series.

FID attn on a 6890N

Posted: Wed Jun 21, 2006 3:11 pm
by chromatographer1
I don't believe changing the attentuation will affect the amplifier setting or change the sensitivity of the detector.

A change in range is required.

best wishes,

Rod

Re: FID attn on a 6890N

Posted: Wed Jun 21, 2006 8:15 pm
by dblux_
AFAIK changing the attenuation affects only the amplitude of analog signal to recorder not the digital signal.
Correct me if I'm wrong.

after checking the manual

Posted: Wed Jun 21, 2006 8:27 pm
by chromatographer1
I checked the documentation on our 6890 and that is correct.

best wishes,

Rod