Recently one of our analysts suggested to increase attenuation, in order to improve sensitivity for the analysis of one residual solvent. We injected a standard, with the following 'table' inserted into the 'runtime' option in the 'instrument actuals' part of the method.

Time Parameter Value
0 Attenuation 0
20 Attenuation 5
22 Attenuation 0

No increase/decrease in area was observed for the two peaks that elute between 20 and 22minutes.

Does anybody know anything about altering attenuation with 6890's?! Any pointers would be really appreciated.
Apologies

We're using an FID on the 6890N series.
I don't believe changing the attentuation will affect the amplifier setting or change the sensitivity of the detector.

A change in range is required.

best wishes,

Rod
AFAIK changing the attenuation affects only the amplitude of analog signal to recorder not the digital signal.
Correct me if I'm wrong.
I checked the documentation on our 6890 and that is correct.

best wishes,

Rod
5 posts Page 1 of 1

Who is online

In total there are 17 users online :: 0 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry