Focus GC with AI3000 and chromquest 5.0 not sending trigger

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

2 posts Page 1 of 1
I have a Focus GC that I have been using for the past 3 years and recently after a couple of months of off time, the autosampler is not sending the trigger to start the injection. It begins the sequence and moves to adjust itself but will sit and "exec prep run events." The problem is that there are no events on the run table and even creating a new method doesn't solve the issue. I have even attempted to configure the autosampler out of the instrument setup but the system will still sit in stand by/prep run mode without ever progressing to the ready to inject state.

Is this a software issue or a mechanical issue? Has anyone ever run into this problem before?
Hello

Go to "Configuration" - "Oven configuration" menu and check :
-Auto prep run
-Auto start
Check them on/off to see if you can change settings.

Regards

Tomasz Kubowicz
2 posts Page 1 of 1

Who is online

In total there are 5 users online :: 0 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 5 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry