- Posts: 4
- Joined: Tue Mar 07, 2017 4:39 pm
I am running the EP method for copper tetraMIBI related substances. Copper tetraMIBI is a Cu complex used in preparation of a pharmaceutical. The LC method is SCX using an Agilent Zorbax.
20µL of the test solution (50 mg/mL) is injected and any impurities found are quantified using the API peak from a 20µL injection of a ref standard, which is a 1000x dilution (in MP) of the test solution. The ref solution is injected first.
(see attached jpg) - Peak #2 is the normal 20µL inj of the ref solution. Peak# 4 is the 20µL inj of the test solution. #4 peak shape is fine (albeit topping out) and elutes at the expected time of about 5 mins. #2 has a shark's fin front and elutes early. #6 is a 100µL inj of the ref soln, it shows slightly more retention. #7 is a 0.1µL injection of the test solution. These chromatograms show that the peak shape and retention deteriorate when lower quantities of API are injected, irrespective of solution concentration.
Anyone have any idea what is going on here?


Thanks