How cross talk is removed in API,Quatro micro mass and thermo LC-MS/MS?
-
- Posts: 13
- Joined: Wed Jan 11, 2006 2:29 am
We have just meet a cross talk problem last week in API 3200. When the method transfer from API 4000 to API 3200, we find IS's peak just after inject the neat solution for X (and we use D9-X as IS), but in API 4000, there's no peak.
Then we change the dull time from 5 ms to 20 ms, then there's no cross talk peak.
Then we change the dull time from 5 ms to 20 ms, then there's no cross talk peak.
The more time I have to think;
The more time I have;
The more time I have;
2 posts
Page 1 of 1
Who is online
In total there are 236 users online :: 1 registered, 0 hidden and 235 guests (based on users active over the past 5 minutes)
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: Bing [Bot] and 235 guests
Most users ever online was 10230 on Thu Dec 04, 2025 5:56 am
Users browsing this forum: Bing [Bot] and 235 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science