We recently received a direct injection residual solvents method (using a split/splitless general purpose glass wool liner), to replace our headspace method as it showed lots of unidentified peaks, thought to be due to sample degradation.
The sample is dissolved in DMSO and tested against a 6-component mixture dissolved in the same solvent. The inlet temperature is set to 140C, while the boiling point of DMSO is 189C. Is this a fundamental flaw? While more, bigger degradation peaks are present at higher inlet temperatures, won't the 'plug' of unvapourised DMSO present in the liner affect reproducibility, sample adsorption, ghost peaks etc?
The problem seems to worsen with increasing number of injections. While initially the sample looked clean, peaks with random retention times and areas showed up as more injections went on.
What can be done to resolve this problem?!