We have a Triplus liquid autosampler hooked up to a Thermo GC Ultra, which we use for fast GC analyses of FAMEs. Unfortunately, about 10% of the samples run in any given sequence do not inject properly. After extensive (and tedious!) testing involving watching autoinjections and adjusting parameters, we have identified the problem as cavitation/bubble formation during the final sample draw-up prior to injection. What we actually see is an air bubble lengthening out from the end of the plunger in the syringe barrel during the sample-draw up step. We have optimized the system as far as possible to prevent this, using the following optimized injection procedure:
Syringe: 10uL 80 mm fixed-needle syringe from SGE with cone tip (23 gauge)
Sample: dissolved in hexane; 50-200uL volume in 1-2mL glass vials closed with 11mm crimp-caps (seal = PTFE/blue silicone).
Injection procedure:
1) 3 x wash with 5uL solvent A (hexane) with draw-up at 10uL/s
2) 2 x rinse with 2uL sample with draw-up at 10uL/s
3) 5 x bubble elimination pumping; 5uL/s strokes with 0.1s delay between strokes
4) Sample draw-up. 1.5uL drawn up at 1uL/s, then 0.5uL expelled at same rate back into sample vial to give final volume of 1uL in syringe. 2s delay before syringe is removed from vial for injection.
5) Inject using rapid injection technique (100uL/s, no delays; using Focusliner containing glass wool).
6) 3 x wash with 5uL solvent B (hexane) with draw-up at 10uL/s to end injection cycle.
We have tried slowing down all the sample and rinse solvent aspiration speeds to 0.1uL/s, but this hasn't helped. We have also changed syringes 4 times, thinking the needle may have been partially blocked or the plunger poorly fitted to the barrel, with no improvement. In all cases the end of the needle is several mm away from the bottom of the vials, so sealing against the bottom of the vial is unlikely. All samples are also particle-free in hexane; we get the same problem with pure standards. Vacuum formation in the vials shouldn't be happening as the solvent volume is 10% or less of the total vial volume.
Why this cavitation is occuring is a complete mystery to me; especially as it only happens on sample draw up, even when the solvent AND sample rinses jsust prior to sample draw-up are fine! And yes, the end of the needle is well below the sample level in all cases; in all cases, re-injection of a suspect sample gives a normal injection and trace.
I would really appreciate any suggestions on how to eliminate this problem. The only solution I can think of is to use a slightly larger gauge needle and/or a needle with a gas-tight design (PTFE in plunger) to minimize possible blockages. Any suggestions would be most welcome.
thanks
Tony