Hi everyone,
I am having problem with the make up gas in the FID of a 6890 Agilent.
After I have cleaned the jet replaced it and installed the column, the FID flame does lit only if the make up gas is set to below 40.
Using a method with higher make up gas flow will cause the flame to go on and off continually.
Today I run some blanks and standards and the flame goes off when the solvent front hit the detector.
Any clue on what can cause this and how I can solve it?

Best Regards,

Alice
Hello

Makeup gas flow is too high. It should be somewhere between 5-40ml/min.
Higher flow will not improve anything and it can blow flame.
Just set it to lower value (25 ml/min) and it will be ok.

Regards

Tomasz Kubowicz
The normal flows in an Agilent FID is that the column flow PLUS the makeup flow should be about 30ml/min ( with the hydrogen flow at 30ml/min). So if you are running a capillary column at 2ml/min, makeup should be 28ml/min. If you are running 1/8 inch packed columns at 30 ml/min, you basically do not need any makeup flow.

Gasman
I NEVER used make up gas more than 40ml/min
The normal flows in an Agilent FID is that the column flow PLUS the makeup flow should be about 30ml/min ( with the hydrogen flow at 30ml/min). So if you are running a capillary column at 2ml/min, makeup should be 28ml/min. If you are running 1/8 inch packed columns at 30 ml/min, you basically do not need any makeup flow.

Gasman
That's what we do also.

Alice,

Try starting at 20ml/min and if it lights and runs ok, inject a standard at 20ml/min, 25ml/min, 30ml/min, 35ml/min and 40ml/min and see what the sensitivity does. You will probably see it begin to drop off by the time you reach 40ml/min. Choose the flow that gives the best sensitivity and stability for your methods.
The past is there to guide us into the future, not to dwell in.
5 posts Page 1 of 1

Who is online

In total there are 15 users online :: 3 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google [Bot], Semrush [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry