We are trying to come up with a LC-MS/MS-method for identifying opiates in urine samples. The method we've created uses a gradient wich starts from 10%meOH/ 90%buffer (10mM NH4Ac) and goes up to 90%meOH/ 10% buffer during 10 minutes, followed by 2 min 100% meOH and another 2 min of the initial gradient for equilibration.
We have a big problem with carryover, mostly with molecules that have the largest logP-values (eg methadone and fentanyl), so the problem is probably related to that.
We have tried almost everything, changing the tubing, the rotor seal, the needle, doing blank injections, washing with methanol and other solvents (like ACN or 2-propanol) in between... We also did a blank gradient, and there was no carryover, so it's obviously a problem with the injector or autosampler.
We have tried different wash/flush-solutions in the gradient, the waters-solution (MeOH-ACN-IPA-water with 1% formic acid) some people on this forum mentioned works best, the carryover is reduced to ca 1%, but then again some peaks (eg. morphine) are really bad-looking, even if the needle is washed with a basic solution before injecting.
I’d be grateful for any ideas how to solve this problem!