Cross-talk Phenomenon in LC-MS/MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Dear all,

Could you all please explain the cross-talk phenomenon in LC-MS/MS to me?

When will it occur? How can we detect it in our method?

In addition, if it occurs, which solutions could be used to troubleshoot?

Many thanks.
Some time ago, the following issue existed in triple quads:

The triple would start measuring transition 1: mass A --> mass B
The first quadrupole would be set to mass A, the final quadrupole to mass B, and you'd measure for some msec.
Then the instrument would change to transition 2: mass C --> mass D
Now, at the instant of change, there may still be ions flowing through the instrument, so it's possible some left-overs of mass A will be passing through the collision cell, and generating fragments, some of which might be able to pass through the final quadrupole even though it's now looking at mass D (especially if you're working on a number of related compounds that will have similar fragments), and there may even be some ions of mass B that are far enough through the final quadrupole, or between it and the detector, that will still show up.
The ability of ions from one transition to be detected in the early part of the next transition's dwell time was called cross-talk. It was minimised by setting an appropriate delay between the two transitions, long enough to clear all ions from the system.

I believe this sort of cross-talk has been completely eliminated in modern instruments.
Some time ago, the following issue existed in triple quads:

The triple would start measuring transition 1: mass A --> mass B
The first quadrupole would be set to mass A, the final quadrupole to mass B, and you'd measure for some msec.
Then the instrument would change to transition 2: mass C --> mass D
Now, at the instant of change, there may still be ions flowing through the instrument, so it's possible some left-overs of mass A will be passing through the collision cell, and generating fragments, some of which might be able to pass through the final quadrupole even though it's now looking at mass D (especially if you're working on a number of related compounds that will have similar fragments), and there may even be some ions of mass B that are far enough through the final quadrupole, or between it and the detector, that will still show up.
The ability of ions from one transition to be detected in the early part of the next transition's dwell time was called cross-talk. It was minimised by setting an appropriate delay between the two transitions, long enough to clear all ions from the system.

I believe this sort of cross-talk has been completely eliminated in modern instruments.
I remember the advertising spec on the ABI 3200 was it could run transitions as short as 5ms without cross-talk. I think they advertised that they reduced the cross-talk by minimizing the internal volume of the collision cell, so that everything gets flushed out faster.
The past is there to guide us into the future, not to dwell in.
3 posts Page 1 of 1

Who is online

In total there are 32 users online :: 3 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry