Waters Alliance 2795 Needle Malfunction Bottom Sensor
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
Hi, I have been using the Waters Alliance 2795 for almost three years now. I was running about 12 samples today, the instrument stopped in on third run, and gave this error message of "Needle Malfunction (Bottom Senor)", I opened the door to the sample plates, I did not observe any jam or mistakes. BTW, we bought this used HPLC unit about 5 years ago, I am not sure if it is still within the warranty. Does anybody now what shall I do with this? Thanks a lot!
-
- Posts: 276
- Joined: Tue Sep 07, 2004 10:05 pm
This is out of the 2795 users guide, at least its a place to start.
Needle failed to detect the Bottom sensor on
reaching the WELL or LOAD position. Check
connections to XYZ mechanism. Remove and
reinstall needle. If error persists, contact Waters
Technical Service.
Needle failed to detect the Bottom sensor on
reaching the WELL or LOAD position. Check
connections to XYZ mechanism. Remove and
reinstall needle. If error persists, contact Waters
Technical Service.
-
- Posts: 2
- Joined: Fri Jan 10, 2014 6:29 pm
Thank you very much! I will start checking on that!This is out of the 2795 users guide, at least its a place to start.
Needle failed to detect the Bottom sensor on
reaching the WELL or LOAD position. Check
connections to XYZ mechanism. Remove and
reinstall needle. If error persists, contact Waters
Technical Service.
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
Some times, turning the system off and on can help.
HTH
Ace
HTH
Ace
-
- Posts: 8
- Joined: Thu Jul 14, 2016 7:54 pm
This is out of the 2795 users guide, at least its a place to start.
Needle failed to detect the Bottom sensor on
reaching the WELL or LOAD position. Check
connections to XYZ mechanism. Remove and
reinstall needle. If error persists, contact Waters
Technical Service.
I also have the sane errir message, may I ask how to check the connections to xyz mechanism and how to remove and re-install needle?
thanks a lot!
5 posts
Page 1 of 1
Who is online
In total there are 11 users online :: 1 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 10 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science