Currently i'm dealing with the dissolution of amlodipine besylate and benazepril hydrochloride capsules,
the mobile phase used is normal TEA buffer with a pH of 3( imean just TEA in Milli-q water and then the pH is adjusted to 3 with OPA)
the column used is 100 x 4.6 mm, 3um hypersil BDS c18
usually the Benazepril peak is eluted at an RT of 6 mins and where as Amlo is eluted at around 8.5 mins, there's no shift in the RT for the first 5 injections of the standard
but when the samples are being injected the amlo peak gradually moves towards the benazepril peak and due to this the brackettings are failing
what could be the reason?
if you say column stabilization .... the RT shift is observed even after the column is stabilised for 2 hours
if you say buffer strength( i mean due to poor buffer strength)... well its a validated method
i'm not understanding where the problem lies
it could be termed as an analytical error if it happens once, but the same problem is seen even after changing the analyst and even the chromatographic system
plz respond immediately