PEAKS DUE TO SYSTEM

Posted: Mon Apr 17, 2006 12:39 pm
by rc_12321
Can any body throw some light on peaks due to the system itself?How to identify them ?Any literature available on this?

Thanks in advance.

Posted: Mon Apr 17, 2006 3:21 pm
by DR
Search posts for the word "Empore" - you'll find a few discussions.

Peaks due to system

Posted: Mon Apr 17, 2006 4:05 pm
by skunked_once
Are you referring to the hardware of the HPLC system? If so, the possibilities are numerous. The most likely sources of peaks are contamination of the mobile phase filters in the solvent bottles and the injector or autosampler. All parts of the flow path in the pump(s), mixer, connecting tubing, and detector could also become contaminated. Most contamination is from the mobile phase and/or sample itself. A good place to look for information on this is LC.GC magazine. Check the archives.

Posted: Mon Apr 17, 2006 5:38 pm