Can any body throw some light on peaks due to the system itself?How to identify them ?Any literature available on this?

Thanks in advance.

Search posts for the word "Empore" - you'll find a few discussions.
Thanks,
DR
Image
Are you referring to the hardware of the HPLC system? If so, the possibilities are numerous. The most likely sources of peaks are contamination of the mobile phase filters in the solvent bottles and the injector or autosampler. All parts of the flow path in the pump(s), mixer, connecting tubing, and detector could also become contaminated. Most contamination is from the mobile phase and/or sample itself. A good place to look for information on this is LC.GC magazine. Check the archives.

-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 18 users online :: 2 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry