http://www.restek.com/catalog/view/3355
That is the connector I use in this setup. You can also use the 0.53 MXT metal transfer line so you don't have to worry as much about breaking the transfer line.
Thanks for your advice. I am investigating the connector. One of my column is HP-5MS is length 30 m - Diam. 0.250 mm - Film 0.25 µm and one of is
DB - VRX Length - 60m / Diam - 0.250 mm / Film - 1.40 um.
You will have to determine for yourself which liner works best for your combination of analytes, solvents, splits and inlet conditions.
The column that i am using for phthalate is HP-5MS is length 30 m - Diam. 0.250 mm - Film 0.25 µm.
My method on this column about 8 min. and that's important for me. I can't extend the duration.
When i try the VOC analysis with this column, the peaks are overloaded and the result is unsuccessful. I wrote here:
viewtopic.php?f=2&t=37761
I need a thicker column for VOC analysis and that's the other column on my hand.
DB - VRX Length - 60m / Diam - 0.250 mm / Film - 1.40 um.