Hi!

I want to find the "volatile organic compounds" with my gcms. My method is sth like that;

oven program: 35°C for 10 minutes
flow: 0.5 mL/min
split: 20/1 split flow: 10mL/min
inlet temperature: 125 °C
inlet: agilent 5190-2195 universal, low pressure drop
column: hp-5ms ultra inert

My solvent and wash solution is methanol.

What i am looking for is; benzen, carbon disulfide, chloroform, diklorometan, ethanol, ethyl acetate, IPA, n-heptane, n-hexane, tetra hydro furan.

But i have a problem. My peaks are nested. How can i get seperate and net peaks?

Image
I would say that you need a different column for your separation. If you can get to subambient oven temperature, you might do a little better. You might also try a thicker film than what you're using right now.

Your analytes are all listed in this one (obtained from http://www.restek.com). Try these conditions:

https://1drv.ms/b/s!AkH-uI0tnY5LdikM7GBX1kX2o9U
At least some of your peaks are broadened by overloading. Increase your split flow to 100 ml/min.

What is the internal diameter of your column ? - your flow rate is unusually slow for the usual column sizes.

Peter
Peter Apps
0.5ml/min isn't too low if you are using a 0.18mm ID column, but the slow flow will broaden the peaks even with that column. I normally run my 0.18 and 0.25 at 0.8 and 1.0ml/min respectively.

For those compounds try the Restek Rxi-624SilMS column in the 20mx0.18x1um film, or the Rtx-502.2 40mx0.18mmx1um film. These work great for volatiles analysis and you can run a high split flow also. The 5 phase with thin film just doesn't work well for such light compounds without sub ambient oven temps.
The past is there to guide us into the future, not to dwell in.
Your analytes are all listed in this one (obtained from http://www.restek.com). Try these conditions:

https://1drv.ms/b/s!AkH-uI0tnY5LdikM7GBX1kX2o9U
I will try this conditions and write here again.
At least some of your peaks are broadened by overloading. Increase your split flow to 100 ml/min. What is the internal diameter of your column ? - your flow rate is unusually slow for the usual column sizes.
I can try increase my split flow to 100 ml/min or try the splitless mod. Which will be more effective?

My column HP-5MS is length 30 m - Diam. 0.250 mm - Film 0.25 µm
Temperature limits from -60 °C to 325 °C (350 °C)
0.5ml/min isn't too low if you are using a 0.18mm ID column, but the slow flow will broaden the peaks even with that column. I normally run my 0.18 and 0.25 at 0.8 and 1.0ml/min respectively.
First i want to try this column for this analysis but i can't do that i will buy a new one that you said.
I can try increase my split flow to 100 ml/min or try the splitless mod. Which will be more effective?
When someone tells you your column is overloaded then stay away from splitless mode.
My column HP-5MS is length 30 m - Diam. 0.250 mm - Film 0.25 µm
Use column with thicker film
You need to do three things:

Increase the flow rate to 2 ml/min (35cm/s linear velocity)

Dilute your samples x 10

Increase your split flow to 200 ml/min with the new column flow.

Peter
Peter Apps
DB5 is not a great residual solvents column. Most recommendations are for 624 or specialized columns.

Usually residual solvents are done on a thicker phase column 1um or so.

You are overloading the column so increase the split ratio.

Use a column flow rate in the ideal portion of the van deempter curve (are you using He or H2?).
...
oven program: 35°C for 10 minutes
...
What i am looking for is; benzen, carbon disulfide, chloroform, diklorometan, ethanol, ethyl acetate, IPA, n-heptane, n-hexane, tetra hydro furan.
...
I doubt you will succeed with this isothermal program.
Hi!

I changed my column. My new column is;

DB - VRX
Length - 60m / Diam - 0.250 mm / Film - 1.40 um
Temperature limit
From -10 to 260

My oven programme is:
60C 2 minutes
15C increase to 180C 0 minutes
45C increse to 225C 5 minutes

Flow 1.3

The peaks that my looking for are coming first 4.7 minute.

My solvent is methanol and blank chromatogram something like that:

Image

On the 4th minute coming a strange peak, before the first my peak Etanol.
How can i clean the peak or how can i develop the method?
Hello
I changed my column. My new column is;

DB - VRX
Length - 60m / Diam - 0.250 mm / Film - 1.40 um
Temperature limit
From -10 to 260

My oven programme is:
60C 2 minutes
15C increase to 180C 0 minutes
45C increse to 225C 5 minutes
Start temperature is too high. Try to start with 35/40

Regards

Tomasz Kubowicz
What makes you think that all your target peaks come out before 4.7 minutes ?

Peter
Peter Apps
What makes you think that all your target peaks come out before 4.7 minutes ?

Peter
rb6banjo share this method conditions about VOC under this topic. The link is here:
https://onedrive.live.com/?authkey=%21A ... ot&o=OneUp
What makes you think that all your target peaks come out before 4.7 minutes ?

Peter
rb6banjo share this method conditions about VOC under this topic. The link is here:
https://onedrive.live.com/?authkey=%21A ... ot&o=OneUp
It's not your column. But bearing in mind mentioned 4.7 min. what is the retention time of tetrahydrofuran in the same conditions ?
What makes you think that all your target peaks come out before 4.7 minutes ?

Peter
rb6banjo share this method conditions about VOC under this topic. The link is here:
https://onedrive.live.com/?authkey=%21A ... ot&o=OneUp
It's not your column. But bearing in mind mentioned 4.7 min. what is the retention time of tetrahydrofuran in the same conditions ?
On my column and my conditions tetrahydrofuran is coming 7.07 min.
16 posts Page 1 of 2

Who is online

In total there are 14 users online :: 0 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry