Lab Contamination

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
I've recently had two instances of a laboratory contamination I'm trying to track down. I have been doing some trace analysis, concentrating down solvent extracts using high purity nitrogen. Once using hexane and a second time using acetone I have found a silicone oil (PDMS) in my extracts and blanks.

Since it occurred using two different solvents, I suspect it may be in my sample concentration step. I am using compressed Ultra High Purity (UHP) grade nitrogen for my blowdown. I have heard that it is possible to pick up contamination from compressed gas sources either when taking the tank down to near empty, or that there can be some bleed from components of the regulator.

Does anybody have any ideas, experienced any similar problems?

Thanks,
Mike
Hi Mike

You do not say whether you are doing HPLC or GC, but I am assuming GC, and that you are detecting by MS.

By far the most likely source of "silicones" is column bleed if you are using a silicone column. Second most likely is septum bleed.

Make an injection of solvent that has not been concentrated by evaporation, if the silicones are still there they are coming form the GC, not from the solvent, gas or laboratory.

Peter
Peter Apps
Actually I should have specified. I am analyzing the extracts by FTIR - evaporating the extracts down on ZnSe crystals and collecting spectra in transimittance mode.

I do have GCMS and HPLC instrumentation in my lab, but in this case FTIR analysis.
Since it occurred using two different solvents, I suspect it may be in my sample concentration step. I am using compressed Ultra High Purity (UHP) grade nitrogen for my blowdown. I have heard that it is possible to pick up contamination from compressed gas sources either when taking the tank down to near empty, or that there can be some bleed from components of the regulator.

Does anybody have any ideas, experienced any similar problems?
The only impurity that normally might increase when cylinder pressure gets low is moisture. The easiest way to check if the cylinder is the problem is to use a different cylinder.

As for the regulator, if you are using one with a rubber diaphragm, that may cause purity problems. Make sure you use one with a stainless-steel diaphragm. (A brass body is fine.)

Oh, and avoid Tygon and similar polymer tubing, they can off-gas plasticizers and such. Copper tubing is a better option. Just make sure it's rated to an appropriate working pressure.

Also, you can use Zero grade nitrogen instead of UHP, if there's a cost or availability issue with the UHP.

How about grease from glass joints?
5 posts Page 1 of 1

Who is online

In total there are 31 users online :: 2 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry