by
lmh » Tue Nov 22, 2016 9:31 am
Following the "Next" link on the Shimadzu tutorial page to which amateur23 posted a link will take you to Shimadzu's description of SIM, which clarifies the way in which it differs from an extracted mass chromatogram.
Extracted mass chromatograms, taken from TIC data, are far more than a sales tool. If you know exactly what you are looking for, you will collect SIM data (assuming you're using a quadrupole instrument) because it's far more sensitive. If you don't know what you're looking for, you will collect scan data. If you then find an interesting ion in the scan data from a complex sample, you'd be insane to integrate the TIC because it might contain hundreds of peaks, mostly coeluting. Instead naturally you will extract a chromatogram of the appropriate mass, which will probably contain only a handful of peaks, and integrate this.
No, it's not as sensitive as SIM, but you can't set up a SIM experiment when you don't know what you're looking for.
This, incidentally, has always been Thermo's argument when selling high-resolution instruments in competition with triple quads: if you're looking for drugs of abuse, MRMs in a triple quad can only look for drugs you know about. If someone comes up with a new drug, you can mine accurate mass data retrospectively. So if you can get your specificity by accurate mass rather than MS^2, your data-set is more useful. I don't completely agree, because MRMs can potentially distinguish isomers, which accurate mass cannot. But Thermo have a point.