im getting a better signal response from my samples in SPLIT mode at 50:1 than I am in SPLITLESS mode.

why!

injecting the same amounts same temp profile and carrier flow rate.

Something to do with the flow!

Splitless only works if you refocus the peaks after the injection. Is your oven temperature below the solent boiling point ?
Also this should be in Gas Chromatography. :)

Sounds to me as though you will soon see peaks matching the retention time of your analyte in your blanks. Might this compound be collecting in your injection port. What is inj port temp, starting oven temp and detector type and temp.

Are you sure that the parameters are set properly for a splitless injection. I remember a similar problem several years ago when a co-worker was getting very poor sensitivity with splitless injection using the default instrument conditions, which included a purge on time setting of 0.0 minutes. After explaining that a purge on time of 0.0 meant the purge vent was on at time of injection, and setting a purge on time of .5 minutes sensitivity improved dramatically.
4 posts Page 1 of 1

Who is online

In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry