I'm currently validating a method and have ran this multiple times now. This is a gradient run, gradient starts at 0.8 min and rises to 11.5 min. changing from Aqueous to Organic. At ~2.5 min i have 2 peaks that come off within about 0.3 min of each other. This has been ran successfully with no splitting multiple times, then we will randomly see the 2 peaks split/ We have see this as a slight shoulder to a full split peak that shows up ~ 0.2 min down field. This does not always occur, and is not column dependent or sample solvent. I say this because we have run it with the same column and same system suitability solution after we have seen splitting and it returns to normal. I cannot seem to find the issue. This has been seen across multiple systems and columns, and again, is very random. We have also ran this and followed it by the same run and we will see splitting. The only true success I have had is re-seating the column and ensuring a very proper column installation. Any help would be great!