Hello all:


I'm new to the lab & would appreciate it if you could help me on the following...


Where can I find guidelines in the USP regarding mobile phase composition adjustment? (e.g. what is the max % organic that can be added to mobile phase for the adjustment of peak retention time)


Thanks in advance.

There was a presentation by William B. Furman from US FDA about " how much can you change an official method before it isn't the official method any more." Based on his presentation and I remembered reading this guideline also from another source, you are allowed to adjust 10% for the ratio of solvent in mobile phase as long as the total composition is 100%. (e.g., 50:50 MeOH:H2O can be adjusted to 45:55 or 55:45).
ntruong

Thanks, ntruong.


Does anybody have a link to a free copy of this presentation by William B. Furman?


Thanks.

I know Mr. Furman has tried to propose these changes to the USP ofiicials to make it official back in 1997, you can contact him at
FURMANW@DDASTL.CDER.FDA.GOV
I hope he is still with the agency.
Good luck,
ntruong

Ntruong:


Thanks again for the info. I will contact him.


I'm thinking that +/- 10% is a huge range for adjustment. For example, if the mobile phase composition is 90:10, it can be adjusted to 99:1 or 89:11 if we were to follow this guideline.


Any ideas?

To the best of my knowledge, there is no official USP guideline at this time. Three roughly comparable proposals have been "test-flown":
  • R. Cox and G. Menon, PharmEuropa, 10 (1998) 58.

    W. B. Furman, J.G. Dorsey, and L.R. Snyder, Pharm. Technol., (June 1998) 58.

    Pharmacopeial Forum, 30(3) (May-June 2004) 1015-1017
The Pharmacopeial Forum reference is a draft of proposed allowable adjustment limits, but as far as I know, those limits have not been incorporated into the 2006 USP. I have no idea whether they will show up in 2007.

Bill Furman retired from the FDA in 1998. His participation in the Pharm Tech article was as a "private individual" (who just happened to share opinions of others within FDA :wink: ).

If I were managing a department, I would write an SOP defining the allowed adjustment levels in your lab, using the articles above as support. Otherwise, you have do justify adjustments on a case-by-case basis.

For proprietary (non-USP) methods, you should define the adjustment limits on the basis of robustness information gathered as part of validation.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
6 posts Page 1 of 1

Who is online

In total there are 8 users online :: 3 registered, 0 hidden and 5 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 5 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry