Question about Shimadzu LCMS-8040 Optimization tool for MRM

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi all.

I´m working with a Shimadzu LCMS-8040 and I was doing a sensitivity test with reserpine. In the S/N test we inject a 1ng/uL of reserpine.

I set the precursor and the product ion as indicated in the protocol, 609,3>195.0. The signal result was about S:1500 and the S/N:1100.

Then used the optimization tool and the precursor ion was adjusted to 609,8>195.0, run the test but the signal decreases a lot. S:150.

I set the precursor ion back to 609,3 again and run the test. The result was S:2572 and the S/N:1287.

I would like to know if someone has seen this, if it is a particular case with this molecule or if it is some resolution of mass issue due to the tuning file.
MRM for reserpine: 609.3 > 195.1 m/z (609.28 > 195.07 m/z)

I think your optimization tool is not working right. Check parameters of optimization tool.

Usually i used auto optimization only at first step (start point) with manual optimization at final step (in most cases i did only manual optimization).
2 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry