I am trying to analyse environmental samples at very low level using HPLC. I faced a problem of noisy base line. The noise is a long-term one, as it is shown in the chromatogram in the following link:
http://us.f13.yahoofs.com/bc/441aeded_2 ... EB9h83AWqd[/img]
The first 100 min of the chromatogram generated at the normal running conditions for my analysis which are: C18 column (250*4.6mm), 70 % ACN/Water mixed by the low-pressure-mixing pump. Flow Rate 1.5 ml/min, DAD set to monitor at 230nm.
105-160 min The column was disconnected as I suspect it to be the cause of the noise. However, the noise continued.
170-200 min I used premixed 70% ACN/Water and put the 2 inlet channels A and B into the same bottle to check if the problem was due to mixing but the problem continued.
> 200 min I stopped the pump and let the DAD on but I got a continuously raising straight line in an angle of about 45 degree.
The large spikes between these periods were due to stopping the pump.
I rally don’t know what may cause such a problem and how I may solve it!
Any advice or comment is very appreciated.
Thank you all in advance.