I have ran a few CoA standards with my polar method (zic-pHILIC ph 9) and I noticed that most compounds (Isobutyryl CoA, Succinyl CoA...) have two very different RTs each. What could be the reason for this?
Ps. My method has 25min and Isobutyryl CoA, for example, gave two peaks: one at 4.7 and other at 10.4!

Any advice is more than appreciated!

Cheers
Questions:

1) What was your mobile phase(s)?
2) What was your sample dissolved in?
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
Hi,

1) A: Ammonium carbonate 20mM
B: ACN

2) I've injected pure standards diluted in MeOH:W.

Thank you!
Sorry, but we need more details. Please describe your gradient and whether or not you generated the solvents by blending a purely aqueous solvent with pure ACN online. Also, what were the percentages of methanol and water in the sample solvent?
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
The gradient was program as decreasing the percentage of B after 2 minutes from 80% to 20% in 15 minutes followed by washing the column at 5% of B for 3 minutes and finally re-equilibrating the column at 80% of B for 5 minutes.

Sample solvent was 50:50 M:W.

Many thanks!
Well, you've got too little salt in the mobile phase, no salt in the sample solvent when you should have some there, too little organic solvent in the sample solvent, the wrong organic solvent in the sample solvent, and a reequilibration time that's at least 6x too short. Other than that, your method's fine...

To be specific:

1) Reequilibration time:
Separation in HILIC involves partitioning into and out of a semi-immobilized layer of water on the surface of the stationary phase. It takes time to set up this layer. Accordingly, it takes longer to reequilibrate a HILIC column than a reversed-phase or ion-exchange column. Try reequilibrating for 30 minutes.

2) If a sample's viscosity is significantly higher than the viscosity of the mobile phase, then the mobile phase won't mix freely with it but will instead tear off streamers of solvent from the periphery of the sample plug as it migrates down the column. These streamers can migrate through the column faster than the rest of the sample. Result: Multiple peaks for a pure solute. Use ACN in your sample solvent instead of MeOH. The concentration of ACN in the sample solvent should be within 10% of the concentration of ACN in the mobile phase. Here, that would be 70% at a minimum.

3) Your starting mobile phase contains 4 mM ammonium carbonate in 80% ACN. Merck doesn't advertise the fact, but at salt concentrations lower than 20 mM overall, its "zwitterionic" ligands have a net negative charge. Accordingly, under the conditions here, your ZIC-pHILIC is a high-capacity cation-exchange material. Now, note that your various CoA's have charges of (-5), four of these from the phosphates in the CoA nucleotide portion and one from the terminal carboxyl- of your succinyl- or butyryl- adduct. Unless you have enough salt in the sample solvent and in the mobile phase, then such analytes will experience pronounced electrostatic repulsion. Peak splitting is common in such cases, and peak separation to the extent that you've observed is not unusual. You need more salt. You also need to have the same salt in the sample solvent and in the mobile phase. Otherwise, the ionizable groups in the CoA compounds will have a variety of counterions. The resulting ion pairs will differ in polarity and will migrate through a HILIC column at different rates, resulting again in multiple peaks (usually connected by a continuum, reflecting the fact that some molecules swap one counterion for another during their migration through the column).
I'd recommend having at least 20 mM ammonium carbonate overall in the starting mobile phase. You could get that by blending ACN with 100 mM ammonium carbonate online in an 8:2 ratio. However, I'd recommend instead that you pre-mix the solvents so that the mobile phase A reservoir actually contains mobile phase A. Same goes for B. Have the same 20 mM salt in both mobile phases, rather than the gradient from 4-16 mM salt that you're using now. Also, the sample solvent should contain at least 20 mM ammonium carbonate overall.

This combination of electrostatic repulsion and HILIC has a name: ERLIC. Here's a link to a paper that discusses it in detail:
http://pubs.acs.org/doi/pdf/10.1021/ac070997p
Look at Fig. 14 for an extreme case of the problem that you've encountered here. You might be better off with 40 mM salt instead of 20 mM, but first let's get everything else right and then assess the importance of that variable.

I don't suppose that you asked Merck for any technical support for your problem?
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
Thank you very much for your answer!

I will try to inject the standards in the solution you suggested and see how it goes. I will also modify the reequilibration time. We've been using this method for more than one year now and it works well for other compounds... Let's see how it goes with the modifications you recommended!

Thanks a lot!
7 posts Page 1 of 1

Who is online

In total there are 387 users online :: 2 registered, 0 hidden and 385 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google Adsense [Bot] and 385 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry