When I try to measure H2 flow out of FID, I found few H2 comes out from FID, Most of H2 flows into the collum and comes out from split purge. Why this happen and what is the problem? Thank you!!
-
- Posts: 200
- Joined: Wed Dec 07, 2005 3:23 pm
Since you are using split to your column, depending on the split used, you will have the splitted part of your flow into the column and then detector, and part to split vent.
Let's say 10 ml/min, split 1:10, you will have 1 ml/min at the end of the column (FID) and 9 ml/min in split vent...
Let's say 10 ml/min, split 1:10, you will have 1 ml/min at the end of the column (FID) and 9 ml/min in split vent...
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
This sounds very very unlikely to me but not quite impossible. If the flame tip is blocked, and you have a very short column installed it is just conceivable that you get back flow from the FID through the column to the inlet and out through the septum purge.
To be completely sure that you are measuring the hydrogen flow to the FID supplied by the detector gas controls of the GC, remove the column from the detector and seal the detector connection where you took the column out. Then measure the hydrogen flow at the detector - it should be around 30 ml / min. While you are about it you should check the make-up and air flows as well. Once you have them adjusted correctly, remove the plug and reconnect the column.
If you cannot get the hydrogen flow that you need by adjusting the controls on the GC, check that you have adequate pressure in your hydrogen supply. Also make sure that the flame tip is not blocked. If you have an Agilent GC the flame tip has to be matched to the column diameter or the column can block the flow of hydrogen.
Good luck. Peter
To be completely sure that you are measuring the hydrogen flow to the FID supplied by the detector gas controls of the GC, remove the column from the detector and seal the detector connection where you took the column out. Then measure the hydrogen flow at the detector - it should be around 30 ml / min. While you are about it you should check the make-up and air flows as well. Once you have them adjusted correctly, remove the plug and reconnect the column.
If you cannot get the hydrogen flow that you need by adjusting the controls on the GC, check that you have adequate pressure in your hydrogen supply. Also make sure that the flame tip is not blocked. If you have an Agilent GC the flame tip has to be matched to the column diameter or the column can block the flow of hydrogen.
Good luck. Peter
Peter Apps
-
- Posts: 586
- Joined: Mon Oct 04, 2004 3:00 am
Be sure you are using the proper adapter to measure flow. Most FIDs come with an adapter that has to be inserted into the top of the detector to be able to measure flow properly. If you just put a flow meter at the top of the detector you will probably not be able to measure the flow properly due to leaks.
4 posts
Page 1 of 1
Who is online
In total there are 112 users online :: 0 registered, 0 hidden and 112 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 112 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 112 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science