We will be working with EPA 625 and 8270, then moving on to 608, 8081,8082 for pesticides and PCBs.
With the Restek method most is going on column, once we can get the Multi Mode inlet, most should be vented out using the solvent vent technique.
We run almost every type of sample, from waste dilutions to drinking waters to soils.
The main push is to reduce solvent usage, and after reviewing the methods, EPA3510C for sep funnel extraction allows to reduce initial volume to 100ml and you use the same 90% reduction in extraction solvent. Method 625.1 was modified so that the initial volume of sample needs to be enough to meet MDL and reporting limit goals and the final extract volume has the same stipulation.
As for needing to clean the inlets, by extracting 100ml sample and injecting 10ul extract you should get the same amount of analytes or crud into the injection port as you would by extracting 1000ml and injecting 1ul. There will probably be some increase in maintenance needed, but if we can achieve a 90% reduction in the methylene chloride it should more than make up for extra liners and gold seals.
I may also experiment with it on the drinking water side to allow for no solvent concentration after SPE, which will save time also on those, if I can dilute up to 10ml instead of concentrating down to 1ml then injecting 10ul. Since that should be fairly clean matrix, it should help with those also. Not really much difference in solvent usage, but that much less that goes up the hood and into the air.