by
Tim » Mon Feb 27, 2006 9:50 am
I would concur with Tom - your approach is too basic to be acceptable to an auditor. In the old days we used to make use of secondary standards, but one too many observations in audits across the company means we now only use primary reference material (either bought in, or extensively tested by a central reference standards group).
A reference material has extensive testing done on it - not just for assay%, but other impurities/degradates, moisture, residual solvents, etc. If you have done similar testing on your "secondary" reference material as was performed on the primary material you tested it against, that would put you in a better position to support using it.
However you also need to prove that your secondary reference material has not degraded over time - how do you know that fridge temperature is sufficient to prevent the standard degrading? One approach would be to retest the secondary reference material against a new batch of primary material after 6 months, 12 months, 18 months, etc. to show there was no change in the purity and no increase in degradates. Ensure you make multiple preparations of all the reference material too so you can show the accuracy and precision of the data you are generating on this ciritical material.
That is not all - you may also need to consider other physical characteristics of the material that may change over time and affect how it behaves when made into a solution.