Accuracy determination using area % for purity determination

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
In our laboratory, we develop and validate many analytical methods to assay various proteins by reversed phase HPLC. Often times, we have disagreement on how to perform the Accuracy part of the validation when the method requires purity determination by Area %. Do you have any suggestions/recommendations on how one could measure this? Is it really possible to check if everything eluted from the column? Anything you can help us will be greatly appreciated.

Some proteins are known to stick to the column and to not elute completely in the first run. This can commonly be seen, when a second blank run is performed (preferentially without injection).

If you are concerned about this, the blank run will reveal if you have a problem. If nothing shows up in the blank run, then you do not have a problem.

If you discover that a substantial amount of your protein(s) elutes in this second run, then more headscratching needs to be done.
2 posts Page 1 of 1

Who is online

In total there are 10 users online :: 2 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry