by
Hornet » Fri Sep 02, 2016 7:48 am
Firstly there was an article written by Davide Balbo but it was erased. Do not have idea why he did İt .
Hi, i deleted it because i missed to read the second chromatogram is a 100 ppm standard, so everything i wrote wasn't answering your question. Plus i was in hurry yesterday and couldn't repost.
I try to tell you how i run this method and what i would change to your instrument conditions to give a better signal, given that your GC is free of leaks and other issues.
I run this method using splitless injection:
splitless time for 1 min then open the valve
liner is single gooseneck with wool
injection volume 2 ul
T injector = 280°C
the main change i would do is the column. The trick is to raise the hydrocarbon hump by shortening the gap between the first eluted peak of interest and the last, for example C10-C40 (unless you NEED to separe them to quantitate single peaks).
Low thermal mass application for TPH from Agilent is a good example of what i'm talking about; it's a different technique but the trick is the same.
My column is 15m and i can go as low as 10 mg/L with the lowest calibration point.
A 60m column will "dilute" your hump with time, it will make hydrocarbon profile hard to integrate or barely visible like your chromatogram.
The baseline rise will always be there, just make sure to acquire a baseline profile everyday and to make your software subtract it from the following chromatogram.
A good baseline is clean from peaks, and when it gets to the latest pleateau it should remain flat, usually it take 1 to 2 blank solvent injection to obtain it.
If everything has gone well, you'll inject your sample and when it's time to integrate you'll just have a flat chromatogram with just the peaks (for diesel) or hump (for oil) ready to be integrated.
In order to optimize splitless injection, since you are using hexane which is good too (i prefer heptane), make your first overn temp ~10°C less than the solvent boiling point to perfectly focus the solvent on the first part of the column. This way your solvent peak will be narrow and won't interfere with C10 integration (which could happen with a short column).
I would try the injector setup i wrote before and this oven program:
60°C for 1 min
20°C/min to 325°C (or whatever is your max allowed temp)
FID has to be at the same temperature as the highest of the overn.
Good work and hope it helps.
Davide