can you help me about ivermectin analysis in sheep sera . Especially extraction method but useless SPE cartridge. Because recovery is too bad. I've a method gokbulut c. add. all. in vet. parasitology 2005 but sample too dirty extraction isn't enough so calibration curve standart 60% and spiked worse than standarts. Please inform me if analysed ivermectin in sera.
hplc user
-
- Posts: 974
- Joined: Tue Jan 11, 2005 11:37 pm
Long, long time ago I worked for the California Veterinary Diagnostic Laboratory System toxicology lab. For some reason, the Midwest Region of AOAC International became the meeting area for people in this field, at least in the U.S. You should try to get in touch with those two organizations. Lots of very helpful people.
Mark Tracy
Senior Chemist
Dionex Corp.
Senior Chemist
Dionex Corp.
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
This may not be easy, but if you give me the current method, I may be able to suggest some improvements.
-
- Posts: 47
- Joined: Wed Feb 09, 2005 7:48 pm
In our lab we determine ivermectin and other macrocyclic lactones in animal tissues by HPLC with FLD detection.
First question is what kind of detector do you use for analysis?
We have some experience with iveremectin in plasma. If you are intersted it it I can send you procedure by e-mail.
First question is what kind of detector do you use for analysis?
We have some experience with iveremectin in plasma. If you are intersted it it I can send you procedure by e-mail.
-
- Posts: 6
- Joined: Thu Feb 16, 2006 3:10 pm
thank you for you reply we use FLD emission 475 ,excitation 475 and we use two different mobil phases firstly acetonitril-methanol ( 66/34), than acetonitril -tetrahydrofuran-water (40-40-20) with the second mobil phase retention time longer than acetonitril methanol 12. mins. my e-mail
g.keskin@alkenet.com
g.keskin@alkenet.com
hplc user
-
- Posts: 6
- Joined: Thu Feb 16, 2006 3:10 pm
thank you for you reply we use FLD emission 475 ,excitation 365 and we use two different mobil phases firstly acetonitril-methanol ( 66/34), than acetonitril -tetrahydrofuran-water (40-40-20) with the second mobil phase retention time longer than acetonitril methanol 12. mins. my e-mail
g.keskin@alkenet.com
g.keskin@alkenet.com
hplc user
-
- Posts: 6
- Joined: Thu Feb 16, 2006 3:10 pm
Pardon Dr_pyrex excitation 365 RegardsIn our lab we determine ivermectin and other macrocyclic lactones in animal tissues by HPLC with FLD detection.
First question is what kind of detector do you use for analysis?
We have some experience with iveremectin in plasma. If you are intersted it it I can send you procedure by e-mail.
hplc user
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Sorry, I was not clear enough. Interferences can be removed with well-thought-out SPE techniques. I am interested in your SPE method. Maybe I can find some ways to imporve it.
9 posts
Page 1 of 1
Who is online
In total there are 6 users online :: 0 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 6 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 6 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science