In series detector set-up
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
5 posts
Page 1 of 1
Heres my situation. Got an 1100 Agilent with mutiple wavelength detector that I need to set up in series with an ELSD detector. We are using Millenium software to run the Agilent. Currently I have the start/stop cable from the Agilent connected to a SAT/IN box on event 1 and the ELSD is connected to channel 1 on the SAT/IN box. We have it set up to run each one individually. Now I need to run the two detectors in series and accumulate the data as two channels so do I run the MWD to channel 2 on SAT/IN box. Help. Thanks.
-
- Posts: 239
- Joined: Tue Nov 09, 2004 3:06 pm
What is your question?
Try to define a new systen that uses both the VWD and the Sat/In. Works fine with Empower.
Alex
Try to define a new systen that uses both the VWD and the Sat/In. Works fine with Empower.
Alex
-
- Posts: 6
- Joined: Wed Feb 08, 2006 4:11 pm
Currently theVWD is connected to an 8 port serial card. The ELSD is connected to the SAT/IN box which then goes to the 8 port serial card. Do I need to run the VWD through the SAT/IN too?
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Using Agilent ChemStation, the Agilent MWD is perfectly compatible with your system as both are 1100s. Its the ELSD which uses the external A/D. With other software I am not familiar.
-
- Posts: 239
- Joined: Tue Nov 09, 2004 3:06 pm
I dont think you should change an cables, as the sysrtem already gets signals from both detectors.
I havn't worked with Millenium, but in Empower there is a subprogramm called system config. In there you can define "chromatographic systems" like autosampler, pump and 1 or 2 detectors. you should already have at least two configurations.
Alex
I havn't worked with Millenium, but in Empower there is a subprogramm called system config. In there you can define "chromatographic systems" like autosampler, pump and 1 or 2 detectors. you should already have at least two configurations.
Alex
5 posts
Page 1 of 1
Who is online
In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 7 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 7 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science