Rob, there have been many "figures of merit" proposed (average resolution, average resolution per minute, . . .). I don't like the various forms of "average" resolution (for general use) because one highly-resolved pair of peaks can result in a high value despite many poorly-resolved peaks. In specific cases, you can get around that by counting only certain problematic peak pairs.
I've generally followed the Lloyd Snyder approach and used the critical resolution (resolution between the worst-separated pair of peaks of interest on the chromatogram) on the grounds that if this "worst-case" pair is well enough separated, all other pairs are, by definition, also well enough separated. A variant on this would be to establish a "target" resolution for each peak pair, and then use the worst-case percentage of target.
For specific problems, I've occasionally played around with applying "fudge factors" to certain peak pairs on the basis of things like disparity in peak sizes, but those get so subjective that I've found it easier to simply use "common sense".